Questions # 11:

While performing duplicate blood gas analyses, a pulmonary function technologist notes that the second instrument gives consistently higher values forPaCO2and PaO2. Which of the following is the most probable cause of higher readings from the second instrument?

Options:

A.

The sample chamber temperature is greater than 37° C


B.

An excessive delay occurred between running the two samples


C.

The PO2and PCO2electrode membranes have protein build-up


D.

The sample chamber has a bacterial contaminate


Expert Solution
Questions # 12:

Which of the following sets of FIF data is most consistent with an extrathoracic airway obstruction?

Options:

A.

Option A


B.

Option B


C.

Option C


D.

Option D


Expert Solution
Questions # 13:

A government agency requires that a pulmonary function technologist establish procedures to minimize the loss of laboratory data. Which of the following should be included in these procedures?

1. Continuous mirroring of hard drives

2. Storage of duplicate test data offsite

3. No Internet connections to the computer storing the database

4. Contracting with a third-party security company

Options:

A.

2, 3, and 4 only


B.

1, 3, and 4 only


C.

1, 2, and 4 only


D.

1, 2, and 3 only


Expert Solution
Questions # 14:

A helium dilution test has just been performed on a patient. The following results are obtained:

FRC 5.0 L

VC 4.0 L

ERV 1.5 L

TLC was calculated to be 6.0 L by plethysmography. From this information, a pulmonary function technologist should conclude that the patient

Options:

A.

Had inadequate intrapulmonary mixing of inspired gas, resulting in an erroneous FRC.


B.

Did not perform the slow vital capacity properly, resulting in too low an FRC by helium dilution.


C.

Was turned into the helium dilution circuit at a lung volume considerably above FRC.


D.

Did not remain in the helium dilution breathing circuit long enough for equilibration.


Expert Solution
Questions # 15:

The following biologic control measurements are obtained:

Which of the following patterns appears in this plot?

Options:

A.

Shift


B.

In control


C.

Drift


D.

Noise


Expert Solution
Questions # 16:

Biological control data are obtained for lung volumes by plethysmography. The following results are obtained:

Question # 16

A pulmonary function technologist should repeat the test instructing the individual to

Options:

A.

Inhale to TV


B.

Pant at 40 to 50 Hz


C.

Pant at 90 to 100 Hz


D.

Inhale to TLC


Expert Solution
Questions # 17:

The following Levy-Jennings charts of control values are obtained:

Which of the following is the correct sequence of actions for a pulmonary function technologist to take?

1. Replace the oxygen electrode.

2. Check the temperature of the measuring chamber.

3. Run protein remover through the blood gas analyzer.

4. Replace the CO2electrode.

Options:

A.

3, then 1


B.

1, then 4


C.

2, then 3


D.

4, then 2


Expert Solution
Questions # 18:

A patient who smokes three packs of cigarettes a day has a DLcoOf 45% of predicted. The patient is referred for exercise (stress) testing. Which of the following should a pulmonary function technologist select to evaluate oxygenation?

Options:

A.

Transcutaneous oxygen electrode


B.

Arterial blood gas analysis


C.

Pulse oximetry


D.

End-tidal oxygen tension


Expert Solution
Questions # 19:

To assure linearity of an oxygen analyzer, calibrate with

Options:

A.

Three test gases within the operating range of the instrument


B.

Air


C.

100% 02


D.

Two test gases within the operating range of the instrument


Expert Solution
Questions # 20:

The following data were obtained from a bronchial challenge test:

Question # 20

APC20of 2 mg/mL was obtained the next day in another laboratory. Which of the following is the most likely explanation for these data?

Options:

A.

The two test results are within day-to-day variability.


B.

Incorrect doses of methacholine were administered.


C.

The spirometer accuracy drifted during the test.


D.

The patient's effort was inconsistent.


Expert Solution