NBRC Registry Examination for Advanced Pulmonary Function Technologists RPFT Question # 17 Topic 2 Discussion

NBRC Registry Examination for Advanced Pulmonary Function Technologists RPFT Question # 17 Topic 2 Discussion

RPFT Exam Topic 2 Question 17 Discussion:
Question #: 17
Topic #: 2

The following Levy-Jennings charts of control values are obtained:

Which of the following is the correct sequence of actions for a pulmonary function technologist to take?

1. Replace the oxygen electrode.

2. Check the temperature of the measuring chamber.

3. Run protein remover through the blood gas analyzer.

4. Replace the CO2electrode.


A.

3, then 1


B.

1, then 4


C.

2, then 3


D.

4, then 2


Get Premium RPFT Questions

Contribute your Thoughts:


Chosen Answer:
This is a voting comment (?). It is better to Upvote an existing comment if you don't have anything to add.