During a third-party audit of a pharmaceutical company (CD9000) having seven COVID-19 vaccination stations in various terminals at a major international airport, you, as audit team leader, interview the CD9000’s Supervisor in Terminal 2.

You: “I see that the temperature of the freezer where you keep the vaccines reads -17.2 °C. I understood that the upper limit for storing vaccines was < -18.0 °C.”

Supervisor: “Yes, I know this. But do not worry, the people who calibrated the temperature sensors said that they have an uncertainty of +/- 1 °C. So, we ensure that the reading remains between -17 °C and -19 °C.”

From the following, select two options that would form your next step.


A.

Ask to see the documented test method to determine if there is additional information on the limits for storage temperature.


B.

Ask to see the calibration report for the measuring equipment.


C.

Raise a non-conformance for storing vaccine at a temperature higher than the maximum value established in the test specification.


D.

Ask to see the competence records of the supervisors of all terminals.


E.

Immediately halt the vaccination process due to the risk to patients.


F.

Considering that the difference of less than 1 degree is insignificant, accept the comment and proceed with the audit.