NBRC Registry Examination for Advanced Pulmonary Function Technologists RPFT Question # 27 Topic 3 Discussion

NBRC Registry Examination for Advanced Pulmonary Function Technologists RPFT Question # 27 Topic 3 Discussion

RPFT Exam Topic 3 Question 27 Discussion:
Question #: 27
Topic #: 3

An isothermal lung analog has the following specifications:

5-L glass bottle

2000 g of steel wool

Density of steel wool 5 g/cm3

Volume of connectors and bulb 100 mi-Repeated FRC measurements in the plethysmograph result in a volume of 5.5 L.

Which of the following should a pulmonary function technologist do?


A.

Remove 500 g of steel wool from the lung analog


B.

Take the plethysmograph out of service


C.

Recalculate the BTPS correction


D.

Begin a biologic control program